The accuracy of dose calculations from a pencil-beam algorithm developed specifically for arc electron beam therapy was evaluated at 10 and 15 MeV. Mid-arc depth-doses were measured for 0' and 90' arcs using 12 and 15 cm radius cylindrical water phantoms. Calculated depth-doses for the 90' arced beams in the build-up region were as much as 3% less than measured values. Calculated values of output (dose per monitor unit) at the depth of the maximum calculated dose were compared with measured values. Isodose contours for a 90' arc were also measured in a 15 cm radius PMMA phantom. At the depth of maximum dose the algorithm predicted doses in the penumbral regions, both with and without collimation, which agreed within a few per cent of measured values. Differences between measurement and calculation are not believed to be clinically significant and are believed to be primarily due to the fact that the algorithm models neither large-angle scattering nor the effects of range straggling on the pencil-beam dose distribution.