The laser-induced breakdown spectroscopy assisted by laser-induced fluorescence (LIBS-LIF) in a two-step process was used to measure the concentration of platinum (Pt) and palladium (Pd) by surface analysis of a solid ore core from the Lac des Iles mine followed by analysis of the same core that was pulverized and compacted. This work focuses mainly on the measurement of Pt since the case of Pd has been extensively discussed in previous work. The excitation of Pt is performed at 235.71 nm with fluorescence emission observed near 269.84 nm. Calibration was performed with synthetic samples prepared from the same ore as the samples studied and the calibration curve shows good linearity in Pt content over several orders of magnitude. A limit of detection (LOD) of approximately 0.15 parts per million (ppm) over an average of 200 laser shots was demonstrated. In contrast, conventional LIBS provides a LOD of about 21 ppm over an average of 200 laser shots due to low signal-to-noise ratio and spectral interference from other elements and does not meet the requirements for estimating the average Pt concentration in the ore. The Pt concentrations obtained using LIBS-LIF on solid ore are generally in good agreement with those obtained in its pulverized and compacted form, as well as with laboratory measurements made by conventional chemical methods. However, the comparison of the results obtained for Pd using LIBS-LIF with the laboratory showed a less satisfactory agreement, probably due to its more inhomogeneous distribution in the ore.