Left-bundle branch area pacing (LBBAP) is a relatively new technique for conduction system pacing. Australian safety and efficacy data is currently lacking. We aim to evaluate the learning curve, medium-term safety, and lead performance in a high-volume Australian setting. We performed a retrospective cohort study of 200 consecutive LBBAP procedures performed by a single operator at two centres between January 2019 and May 2020. Left bundle branch area pacing was performed predominantly via left subclavian access using a 69 cm Medtronic SelectSecure 3830 pacing lead and a preformed non-steerable C315-His sheath. Procedural success was defined as evidence of left septal or left bundle branch area capture as evidenced by a right bundle branch block-like paced morphology. Procedural characteristics, and follow-up (including lead performance) data were collected. Procedural efficiency over time, as well as safety data, were collected. Median age was 78.26 years (interquartile range [IQR] 71-85), 37% were female. Atrial fibrillation was present in 22%. The left ventricular ejection fraction <50% in 24%, atrioventricular (AV) block was present in 43.5%, left bundle branch block in 22.5% and right bundle branch block in 24.5%. Acute procedural success was 91.5%. Implant threshold was 0.6V @ 0.5 ms, and 0.75V @ 0.5 ms at 11.9 months of follow-up. The QRS was significant reduced (baseline 134 ms vs implant 114 ms, p<0.001) There was a reduction in procedural time and X-ray dose over the course of the study. There were no complications specific toLBBAP. LBBAP appears to be a safe and effective pacing strategy. The QRS duration was significantly reduced compared to baseline. There appears to be an early learning curve with LBBAP.