This study aims to evaluate the optical response dependence of the PAKAG polymer gel dosimeter on photon energy and dose rate. The produced gel dosimeters were irradiated using a Varian CL 21EX medical linear accelerator with delivered doses of 0, 2, 4, 6, 8, and 10 Gy. To examine the response dependence on the delivered dose rate, dose rates of 50, 100, 200, and 350 cGy min−1 were investigated. Additionally, two incident beam qualities of 6 and 18 MV were examined to study the response dependence on the incident beam energy. The irradiated polymer gel dosimeters were readout using a UV–vis spectrophotometer in the 300 to 800 nm scan range. The results reveal that a wide variation in dose rate (50–350 cGy.min−1) influences the absorbance-dose response and the sensitivity of PAKAG gel. However, smaller variations did not show a significant effect on the response. Furthermore, the response changed insignificantly with beam quality for investigated energies. It was concluded that the optical reading response of the PAKAG polymer gel dosimeter is satisfactorily independent of external parameters, including dose rate and incident beam quality.