Purpose The main objective is to obtain a method to validate the dosimetry of the VMAT treatment plans when these have several arcs with different isocenters and those arcs overlap. The main interest is to check that no hot or cold dose regions are produced when the treatment is delivered. Methods The dosimetry was calculated with the Eclipse radiotherapy treatment planning system (TPS) version 13.7 for the patient and for the phantom that was going to be used to check the dosimetry. The treatment was irradiated with a Varian DHX linear accelerator and the dose was measured with the dose Octavius 4D phantom and the Octavius detector 729. The fields were measured individually and the phantom was not shifted when it was irradiated. The measured dose was converted to a tiff image with the software VeriSoft. The calculated dose in the TPS was also converted to a tiff image. A composed dose image was created from the individual dose images with the software imageJ. The dose images where shifted when the final dose image was composed by addition of the individual image dose values. The tiff image of the whole treatment obtained from the TPS was compared to this image. Results Images were compared with profiles on the obtained images. The average difference in the ROI was a 3.33%. The maximum difference found was a 23%. A cold spot was created in the actual treatment. It appeared in one edge of the treatment dose calculation of the planning system. There was no arc overlapping at that position of the treatment. Conclusions According to the results, this is a reliable method to check whether an arc treatment with several isocenters can be delivered correctly by the treatment linear accelerator.