Non-contact capacitive coupling ECG (cECG) is considered to have a broad application prospect for its comfort and non-invasive. As an important part of cECG transmission path, the characteristics of dielectric materials significantly affect the signal quality. This paper studies the influence on the dielectric properties of coupling capacitance, human body-electrode impedance, and cECG signal quality from the perspectives of cotton fabric humidity and thickness and gives suggestions on the application of cECG. Firstly, the effects of fabric humidity and thickness on human body-electrode impedance are analyzed theoretically. Then the human body-electrode impedance and cECG signal quality index (SQI) are recorded and calculated to evaluate how thickness and humidity affect the sensing performance of cECG. The results show that influenced by the fabric thickness and humidity, the impedance values reveal observable changes varying from 5.6 GΩ (thickness: 1.00 mm, moisture regain 3.2%) to 50.4 MΩ (thickness: 0.18mm, moisture regain 23.2%) at 0.5 Hz, and the impedance change will cause the impedance mismatch between the two electrodes, resulting in poor signal quality. According to the trend of SQI, the recommended moisture regain under different thicknesses is given. When the fabric thickness is 0.18 mm, 0.35 mm, and 1.00 mm, the moisture regain should be higher than 8.5%, 14.6%, and 23.2%, respectively so as to obtain a relatively high-quality cECG signal. The design parameters discussed in this paper could provide a supporting evidence for the design of efficient non-contact cECG systems in the future.