PurposeTo characterize the accuracy of noncoplanar image guided radiation therapy with the Varian Edge radiosurgery system for intracranial stereotactic radiosurgery (SRS) treatments by assessing the accuracy of kV/kV orthogonal pair registration with Optical Surface Monitoring System (OSMS) monitoring relative to cone beam computed tomography (CT). Methods and materialsA Computerized Imaging Reference System head phantom and Encompass SRS Immobilization System were used to determine collision-free space for orthogonal image pairs (kV/kV) for couch rotations (CRs) of 45°, 30°, 15°, 345°, 330°, and 315°. Couch-induced shifts were measured using kV/kV orthogonal image pairs, OSMS, and cone beam CT. The kV/kV image pairs and OSMS localization accuracy was also assessed with respect to cone beam CT. ResultsMean orthogonal image pair differences for 315°, 330°, 345°, 15°, 30°, and 45° CRs were ≤±0.60 mm and ±0.37°. OSMS localization accuracy was ≤±0.25 mm and ±0.20°. Correspondingly, kV/kV localization accuracy was ≤±0.30 mm and ±0.5°. Shift differences for various image pairs at all CRs were ≤±1.10 mm and ±0.7°. Cone beam CT deviation was 0.10 mm and 0.00° without patient motion or CR. ConclusionBased on our study, CR-induced shifts with the Varian Edge radiosurgery system will not produce noticeable dosimetric effects for SRS treatments. Thus, replacing cone beam CT with orthogonal kV/kV pairs coupled with OSMS at the treatment couch angle could reduce the number of cone beam CT scans that are acquired during a standard SRS treatment while providing an accurate and safe treatment with negligible dosimetric effects on the treatment plan.