An image-quality CT phantom was scanned with three different 3D X-ray imaging guidance devices in the operating theatre: O-Arm, Loop-X, and Airo TruCT. Default acquisition and reconstruction parameters for lumbar spine procedures were used on each device. The tube current was set to a dose level of around 27 mGy. A task-based image quality assessment was performed by calculating the noise power spectrum (NPS) and task transfer function (TTF). A detectability index (d’) was calculated for three simulated bone lesions. The noise magnitude of the O-Arm was higher than the Airo TruCT, and the Loop-X had higher noise than the Airo TruCT. The highest average NPS frequency was for the O-Arm images, and the lowest was for the Loop-X. The TTFs at 50% values were similar for the Airo TruCT and Loop-X devices. Compared to Airo TruCT, the TTF at 50% value increased with the O-Arm by 53.12% and 41.20% for the Teflon and Delrin inserts, respectively. Compared to Airo TruCT, the d’ value was lower with Loop-X by −26.73%, −27.02%, and −23.95% for lytic lesions, sclerotic lesions, and high-density bone, respectively. Each 3D-imaging spine surgery guidance device has its own strengths and weaknesses in terms of image quality. Cone-beam CT systems apparently offer the best compromise between noise and spatial resolution for spine surgery.