In this paper, a comprehensive hybrid K-edge/XRF densitometer (HKED) device model is constructed using MCNP simulation. After the modeling process, a systematic simulation study is conducted to analyze the physical parameters and material selection of KED and XRF. The simulation results reveal that the optimal parameters for the X-ray tube are an X-ray source voltage of 160 kV and a 1 mm Fe filter. The sample should be placed in a vial with an inner diameter of 1.4 cm and an outer diameter of 2 cm. For the KED technique, the determined main parameters are a 1.9 cm Fe filter rod and an inner diameter of 0.08 cm for the collimator. For the XRF technique, the determined main parameters are a 0.01 cm Gd filter and an inner diameter of 0.3 cm for the collimator, with a detector angle of 150°. After selecting appropriate parameters, the average calibration factor Δμ of the KED technique was found to be 3.301 cm2 g−1, with a relative standard deviation (RSD) of 3.36%. Additionally, the comparison between the simulated and calculated values of uranium concentration revealed a minimum measurement error of 0.4%. The minimum detection concentration of KED for uranium solutions is approximately 1 g/L. For plutonium solutions ranging from 0.5 to 20 g/L, linear fitting of the Ka1 net peak area and plutonium concentration showed a coefficient of determination (R2) of 0.999. The detection limit of XRF for plutonium measurement was 2.33✕10−4 g/L. The linear fitting coefficients (R2) of uranium concentration versus K-edge transmission rate and plutonium concentration versus Ka1 net peak area for the hybrid technique in measuring uranium-plutonium mixed solutions are determined as 0.999 and 0.996, respectively, demonstrating the response relationship of the HKED device to uranium and plutonium under different concentrations.
Read full abstract