During computed tomography (CT), a large amount of ionizing radiation is emitted to ensure high quality of the obtained radiological image. This study measured the dose distribution around the CT scanner and the exposure of people staying near the CT scanner during the examination. The measurements used an anthropomorphic phantom to assess human exposure to ionizing radiation. The probability of inducing leukemia and other cancers as a result of absorbing doses recorded around the CT device was also calculated. The highest exposure to scattered radiation in the proximity of the CT scanner is recorded at the gantry of the tomograph, i.e., 55.7 μGy, and the lowest, below lower detection limit of 6 μGy at the end of the diagnostic table. The whole-body detector placed on the anthropomorphic phantom located at the diagnostic table right next to the CT gantry recorded 59.5 μSv and at the end of the table 1.5 μSv. The average doses to the lenses in these locations were: 32.1 μSv and 2.9 μSv, respectively. The probability of induction of leukemia or other types of cancer is low, but the need for people to stay in the examination room during a CT examination should be limited to the necessary minimum. Int J Occup Med Environ Health. 2024;37(3):326-34.