Introduction: With coronary artery disease being the leading cause of death worldwide, it is essential to maximize the efficiency of its diagnosis, which routinely involves SPECT myocardial perfusion imaging (MPI). Using phantom studies, we investigated the influence of time‐reduced SPECT MPI acquisitions on selected measurements of the left ventricle perfusion. Materials & Methods: A thorax phantom with a normal heart was scanned using Philips' Forte JS SPECT system using a clinical cardiac acquisition protocol. Four data sets with 5, 10, 15, and 20 seconds per projection (spp) were acquired concurrently in a single scan. Three additional acquisitions were performed using 3 different numbers of projections: 64, 32, and 16. Reconstructions were performed with Philips' Astonish software and our in‐house reconstruction software (MIRG). Phantom myocardial wall thickness (MWT) and voxel by voxel count comparisons were made on all reconstructed data sets.Results: No significant differences were observed in MWT between reconstructions with different times per projection. For Astonish reconstructions, the MWT in the 20spp, 15spp, 10spp and 5spp were 12.2±1.2mm, 12.6±0.9mm, 13.1±1.6mm, and 11.9±1.2mm, respectively. The average differences in counts per voxel within the heart relative to the gold standard (20spp) were 2.8%, 13.8%, and 15.9% for 15spp, 10spp, and 5spp, respectively. Similar results were observed for MIRG reconstructions and reduced number of projection protocols. Conclusions: Reducing the time of acquisition can be done without comprising accuracy of MWT measurements, but caution should be used if relative quantitation is to be considered, such as in perfusion defect analysis.