Modern radiotherapy techniques often deliver small radiation fields. In this work, a practical Electron Paramagnetic Resonance (EPR) dosimetry protocol is adapted and applied to measure output factors (OF) in small fields of a 6 MV radiotherapy system. Correction factors and uncertainties are presented and OFs are compared to the values obtained by following TRS-483 using an ionization chamber (IC). Irradiations were performed at 10cm depth inside a water phantom positioned at 90cm source to surface distance with a 6 MV flattening filter free photon beam of a Halcyon radiotherapy system. OFs for different nominal field sizes (1×1, 2×2, 3×3, 4×4, normalized to 10×10 cm2 ) were determined with a PinPoint 3D (PTW 31022) IC following TRS-483 as well as with alanine pellets with a diameter of 4mm and a height of 2.4mm. EPR readout was performed with a benchtop X-band spectrometer. Correction factors due to volume averaging and due to positional uncertainties were derived from 2D film measurements. OFs obtained from both dosimeter types agreed within 0.7% after applying corrections for the volume averaging effect. For the used alanine pellets, volume averaging correction factors of 1.030(2) for the 1×1 cm2 field and <1.002 for the larger field sizes were determined. The correction factor for positional uncertainties of 1mm was in the order of 1.018 for the 1×1 cm2 field. Combined relative standard uncertainties uc for the OFs resulting from alanine measurements were estimated to be below 1.5% for all field sizes. For IC measurements, uc was estimated to be below 1.0%. A practical EPR dosimetry protocol is adaptable for precisely measuring OFs in small fields down to 1×1cm2 . It is recommended to consider the effect of positional uncertainties for field sizes <2×2cm2 .
Read full abstract