Novel CT reconstruction techniques strive to maintain image quality and processing efficiency. The purpose of this study is to investigate the impact of a newer hybrid iterative reconstruction technique, Adaptive Statistical Iterative Reconstruction-V (ASIR-V), in combination with various CT scan parameters on the semi-automated quantification using various lung nodules. A chest phantom embedded with eight spherical objects was scanned using varying CT parameters such as tube current and ASIR-V levels. We calculated absolute percentage error (APE) and mean APE (MAPE) using differences between the semi-automated measured diameters and known dimensions. Predictive variables were assessed using a multivariable general linear model. The linear regression slope coefficients (β) were reported to demonstrate effect size and directionality. The APE of the semi-automated measured diameters was higher in ground-glass than solid nodules (β=9.000, p<0.001). APE had an inverse relationship with nodule diameter (mm; β=-3.499, p<0.001) and tube current (mA; β=-0.006, p<0.001). MAPE did not vary based on the ASIR-V level (range: 5.7%-13.1%). Error is dominated by nodule characteristics with a small effect of tube current. Regardless of phantom size, nodule size accuracy is not affected by tube voltage or ASIR-V level, maintaining accuracy while maximizing radiation dose reduction.