Background: Inadequate removal of molecules between 5 and 50 KDa may cause long-term complication in chronic hemodialysis. Medium cut-off (MCO) is a new class of membranes with enhanced sieving properties and negligible albumin loss. MCO membrane makes it possible to perform expanded hemodialysis (HDx), a technique based on high internal filtration (IF).The present study is designed to quantify IF in 2 MCO dialyzers (Theranova 400 and 500, Baxter, Deerfield, USA) using a nuclear imaging technique previously validated. Methods: Blood and dialysate compartment pressure drop along with transmembrane pressure; they were measured in a closed in vitro circuit with human blood (blood flow [Q<sub>B</sub>] = 300 and 400 mL/min; dialysate flow 500 mL/min; net ultrafiltration rate 0 mL/min). A non-diffusible marker molecule (albumin macro-aggregates labeled with <sup>99</sup>Tc metastable) was injected in the blood compartment and nuclear emission was recorded by a gamma camera. Relative variations in the concentration of the marker molecule along the length of the filter were used to calculate local cross filtration. Results: Based on marker concentration profiles, IF was estimated. For Theranova 400, IF were 29.7 and 41.6 mL/min for Q<sub>B</sub> of 300 and 400 mL/min. For Theranova 500, IF were 31.6 and 53.1 mL/min for Q<sub>B</sub> of 300 and 400 mL/min respectively. Conclusions: MCO membrane provides significant amounts of IF due to the particular combination between hydraulic permeability of the membrane and reduced inner diameter of the fibers. High IF combined with enhanced sieving profile of MCO membrane leads to improved removal of a wider spectrum of uremia retention molecules in HDx, without requiring complex equipment.