The aim of this study was to retrospectively assess patient treatment log files to determine an optimal imaging frequency for Intracranial Stereotactic Cyberknife treatments and thus, reducing the imaging dose received by the patient to ALARA while maintaining acceptable delivery accuracy. Image guided radiotherapy (IGRT) and stereotactic radiotherapy (SRT) have become fundamental treatment modalities in radiotherapy for ensuring accurate dose distribution to the intended target. To assess intra-fractional motion, frequent tracking can be employed to accurately track target motion during treatment. If an ionising radiation modality is used, the additional imaging dose received by the patient should be optimised.Patient entrance surface dose (ESD) was measured utilising solid water phantoms to mimic patient scatter for exposures used clinically. Imaging dose was quantified by applying ESD measurements to retrospective data for a cohort of 100 fractions and analysing patient position at each acquired image pair. Patient movement was established by the difference between two consecutive images for the duration of the treatment fraction. Patient movement along with plan robustness was used to determine imaging frequency. ESD was measured to be 0.24 mGy per image pair. In comparison to an original average imaging frequency of 43 s over all patients, an imaging frequency of 64 and 15 s has been suggested for Meningiomas (MEN)/Acoustic Neuromas (AN) and Trigeminal Neuralgia (TGN) respectively. By optimising the imaging frequency, the imaging dose received by patients may increase by up to 150% (TGN) or decrease by up to 20% (MEN/AN) while adequately tracking patient movement.