A horizontal-type 6 MV Pelletron tandem accelerator was installed at the University of Tsukuba, Japan, in September 2014. The first test of the multi-nuclide accelerator mass spectrometry system was performed in March 2016, after two years of construction. This is the first report of its performance. The particle detection system was designed for high-sensitivity detection of 10Be, 14C, 26Al, 36Cl, 41Ca, and 129I. A five-electrode ΔE-E gas-ionization detector was installed as the end station of the detection system. The system background for 14C measurements, with the 4+ charge state at 5.0 MV, has a 14C/12C ratio of ∼2.4 × 10−16. 10Be, 26Al, 36Cl, 41Ca, and 129I have also been successfully detected. The 41Ca/Ca background is ∼3 × 10−15 as a result of improved isobaric separation techniques.