BackgroundMedical exposure from diagnostic X-ray machines represents the primary artificial source of human exposure to low ionizing radiation. Knee X-rays are a standard examination to diagnose multiple conditions ranging from traumatic injuries, degeneration, and cancer. In Saudi Arabia, there is a lack of available data and pertinent information regarding the assessment of the patient entrance surface dose (ESD) from knee X-rays. This study aims to estimate the ESD for adult patients undergoing knee X-ray examinations at four major medical centres in Saudi Arabia. Methods and materialsThe ESD was estimated indirectly using X-ray exposure factors for patients and directly using optically stimulated luminescence (OSL). The results were compared with previous studies conducted nationally and internationally and with international reference levels. ResultsThe results of this research showed that the mean patient ESD was 1.4 ± 0.37 mGy using the direct method and 1.12 ± 0.20 mGy using the indirect method. The estimated mean ESD values were found to be much higher than the mean ESD reference values, emphasizing the crucial requirement for optimizing radiation doses in conventional radiology to improve patient safety and diagnostic effectiveness. ConclusionThis confirms the requirement for reducing patients’ doses to the tolerable levels recommended by many international radiation protection commissions.