Magnetic resonance imaging (MRI) using 1.5T or 3.0T systems is routinely employed for assessing wrist pathology; however, due to off-resonance artifacts and high power deposition, these high-field systems have drawbacks for real-time (RT) imaging of the moving wrist. Recently, high-performance 0.55T MRI systems have become available. In this proof-of-concept study, we tested the hypothesis that RT-MRI during continuous, active, and uninterrupted wrist motion is feasible with a high-performance 0.55T system at temporal resolutions below 100 ms and that the resulting images provide visualization of tissues commonly interrogated for assessing dynamic wrist instability. Participants were scanned during uninterrupted wrist radial-ulnar deviation and clenched fist maneuvers. Resulting images (nominal temporal resolution of 12.7-164.6 ms per image) were assessed for image quality. Feasibility of static MRI to supplement RT-MRI acquisition was also tested. The RT images with temporal resolutions < 100 ms demonstrated low distortion and image artifacts, and higher reader assessment scores. Static MRI scans showed the ability to assess anatomical structures of interest in the wrist. RT-MRI of the wrist at a high temporal resolution, coupled with static MRI, is feasible with a high-performance 0.55T system, and may enable improved assessment of wrist dynamic dysfunction and instability. Real-time MRI of the moving wrist is feasible with high-performance 0.55T and may improve the evaluation of dynamic dysfunction of the wrist.