Background/Objectives: Head Computed Tomography (CT) is an essential diagnostic tool for identifying brain pathologies and visualizing blood vessels. However, CT exposes patients to ionizing radiation, making it necessary to establish local diagnostic reference levels (DRLs) to ensure patient safety. This study aimed to establish DRLs for head CT scans and assess the influence of patient characteristics on radiation dose. Methods: A retrospective analysis was conducted on 2043 non-contrast and 488 contrast-enhanced head CT scans performed between 1 July 2023 and 31 March 2024 using a SIEMENS SOMATOM Definition Edge machine. Computed Tomography Dose Index (CTDIvol) and Dose-Length Product (DLP) values were analyzed, with DRLs set at the 75th percentile. The influence of gender, height, and weight on radiation dose was also evaluated. Results: The DRL for both non-contrast and contrast-enhanced scans was 58.18 mGy for CTDIvol and 1018.11 mGy·cm for DLP per acquisition. Total DLP was 2046.09 mGy·cm for contrast-enhanced and 1027.99 mGy·cm for non-contrast scans. No significant correlation was found between patient characteristics and radiation dose, allowing for a uniform DRL to be established. Conclusions: Uniform DRLs were successfully established for head CT scans, ensuring safe radiation doses for both non-contrast and contrast-enhanced studies. The lack of correlation between patient-specific factors and dose supports the use of standardized DRLs, contributing to optimized radiation safety in head CT diagnostics.