Low-pressure gas chromatography–mass spectrometry (LP-GC–MS) using a quadrupole MS instrument was further optimized and evaluated for the fast analysis of multiple pesticide residues in food crops. Performance of two different LP-GC–MS column configurations was compared in various experiments, including ruggedness tests with repeated injections of pesticides in matrix extracts. The tested column configurations employed the same 3 m × 0.15 mm i.d. restriction capillary at the inlet end, but different analytical columns attached to the vacuum: (A) a 10 m × 0.53 mm i.d., 1 μm film thickness RTX-5 Sil MS column; and (B) a 10 m × 0.25 mm i.d., 0.25 μm film thickness DB-5MS column. Under the optimized conditions (compromise between speed and sensitivity), the narrower analytical column with a thinner film provided slightly (<1.1-fold) faster analysis of <5.5 min separation times and somewhat greater separation efficiency. However, lower detection limits for most of the tested pesticides in real extracts were achieved using the mega-bore configuration, which also provided significantly greater ruggedness of the analysis (long-term repeatability of analyte peak intensities, shapes, and retention times). Additionally, the effect of the increasing injection volume (1–5 μl) on analyte signal-to-noise ratios was evaluated. For the majority of the tested analyte–matrix combinations, the increase in sensitivity caused by a larger injection did not translate in the same gain in analyte detectability. Considering the costs and benefits, the injection volume of 2–3 μl was optimal for detectability of the majority of 57 selected pesticides in apple, carrot, lettuce, and wheat extracts.