In this paper we present a further step in the implementation of a physical phantom designed to generate sets of "true" independent reference data as requested by TG-186, intending to address and mitigate the scarcity of experimental studies on brachytherapy (BT) validation in heterogeneous media. To achieve this, we incorporated well-known heterogeneous materials into the phantom in order to perform measurements of 125I dose distribution. The work aims to experimentally validate Monte Carlo (MC) calculations based on MBDCA and determine the conversion factors from LiF response to absorbed dose in different media, using cavity theory. The physical phantom was adjusted to incorporate tissue equivalent materials, such as: adipose tissue, bone, breast and lung with varying thickness. MC calculations were performed using MCNP6.2 code to calculate the absorbed dose in the LiF and the dose conversion factors (DCF). The proposed heterogeneous phantom associated with the experimental procedure carried out in this work yielded accurate dose data that enabled the conversion of the LiF responses into absorbed dose to medium. The results showed a maximum uncertainty of 6.92 % (k = 1), which may be considered excellent for dosimetry with low-energy BT sources. The presented heterogeneous phantom achieves the required precision in dose evaluations due to its easy reproducibility in the experimental setup. The obtained results support the dose conversion methodology for all evaluated media. The experimental validation of the DCF in different media holds great significance for clinical procedures, as it can be applied to other tissues, including water, which remains a widely utilized reference medium in clinical practice.