We investigated the performances of two computed tomography (CT) systems produced by the same manufacturers (Somatom Flash and Edge Siemens) with different detector technologies (Ultrafast Ceramic and Stellar) and different generation of iterative reconstruction (IR) algorithms (SAFIRE and ADMIRE). A homemade phantom was scanned and the images were reconstructed with filtered back-projection (FBP) and IR algorithms. In terms of image quality, the performances of the systems were checked using the low-contrast detectability, evaluated by a Channelized Hotelling Observer (CHO), and the noise power spectrum (NPS). The analysis with CHO showed the best performance of Edge respect to Flash system for both FBP and IR algorithms. This better behavior, which reaches 20%, has been ascribed to the Stellar detector. From the NPS analysis, the noise reduction due to Stellar detector was 57%, moreover ADMIRE algorithm preserves a more traditional CT image texture appearance versus SAFIRE due to a lower NPS peak shift.