Ventricular bipolar voltage values <0.5 and <1.0/1.5 mV (epi- and endocardium) correlating with dense scar and border zone, respectively, were established using a 3.5-mm tip catheter. Novel microelectrode catheters promise improved mapping resolution; however, whether standard voltage criteria apply to catheters with smaller electrode size and interelectrode distance remains unclear. The purpose of this study was to determine whether traditional bipolar voltage criteria for scar apply during substrate mapping with a microelectrode catheter. Paired bipolar and microbipolar voltage values were acquired from control swine (n = 2) using the microelectrode catheter and assessed for systemic differences. In a postinfarction swine model (n = 6), scar characteristics were compared between the bipolar maps and microbipolar maps using both standard and adjusted voltage criteria derived from the control animals. In control swine, although 5th percentile values for bipolar and microbipolar voltage were similar (1.12 vs 1.22 mV [leftventricular (LV) endo]; 0.88 mV vs 0.98 mV [epi]), median values were significantly greater when acquired by microbipolar electrodes (3.60 vs 6.76 mV, P = .002 [LV endo]; 2.61 vs 2.72 mV, P = .02 [epi]). Microbipolar values were systematically larger by 2.0× and 1.4× in the LV endocardium and epicardium, respectively. Application of standard voltage values to microbipolar maps in postinfarct swine underestimated scar area by approximately 41% in the LV endocardium (13.7 vs 33.4 cm2, P = .004). Bipolar voltage values acquired from microelectrodes are systemically larger than those acquired from standard catheters. New reference values should be established for these novel catheters.