BackgroundWe retrospectively assessed two types of sutureless screw-in left ventricular (LV) leads (steroid eluting vs. non-steroid eluting) in cardiac resynchronization therapy (CRT) implantation with regards to their electrical performance.MethodsBetween March 2008 and May 2014 an epicardial LV lead was implanted in 32 patients after failed transvenous LV lead placement using a left-sided lateral minithoracotomy or video-assisted thoracoscopy (mean age 64 ± 9 years). Patients were divided into two groups according to the type of implanted lead. Steroid eluting (SE) group: 21 patients (Myodex™ 1084 T; St. Jude Medical) and non-steroid eluting (NSE) group: 11 patients (MyoPore® 511,212; Greatbatch Medical).ResultsAll epicardial leads could be placed successfully, without any intraoperative complications or mortality. With regard to the implanted lead following results were observed: sensing (mV): SE 8.8 ± 6.1 vs. NSE 10.1 ± 5.3 (p = 0.380); pacing threshold (V@0.5 ms): SE 1.0 ± 0.5 vs. NSE 0.9 ± 0.5 (p = 0.668); impedance (ohms): SE 687 ± 236 vs. NSE 790 ± 331 (p = 0.162). At the follow-up (2.6 ± 1.9 years) the following results were seen: sensing (mV): SE 8.7 ± 5.0 vs. NSE 11.2 ± 6.6 (p = 0.241), pacing threshold (V@0.5 ms): SE 1.4 ± 0.5 vs. NSE 1.0 ± 0.3 (p = 0.035), impedance (ohms): SE 381 ± 95 vs. NSE 434 ± 88 (p = 0.129).ConclusionsBased on the results no strong differences have been found between the both types of epicardial LV leads (steroid eluting vs. non-steroid eluting) in CRT implantation in short- and midterm.