An advanced instrument employing Laser Photolysis-Laser Induced Fluorescence (LP-LIF) technique has been developed for the rapid and accurate measurement of atmospheric OH reactivity (kOH). A reactive flow tube with a homogeneous flow field and a highly sensitive OH radical detection unit was developed to improve the detection performance of the system. The detection limit for kOH was found to be 0.19 s⁻¹ with a high time resolution of 10 s. Implementing a new digital algorithm for signal processing helps improving the time resolution to 1 s while keeping a low detection limit at 0.21 s⁻¹. To validate the accuracy of the system, the reactions of OH with single gases (CH4 and CO) and mixed gases (CH4, CO and NO) were assessed. The measurement values closely matched the expected values reported by the International Union of Pure and Applied Chemistry (IUPAC). The potential interference effects of NO, O3, and H2O were extensively investigated, thereby enhancing the instrument's applicability in complex pollution environments. Field observations were successfully conducted in Hefei, Nanjing, and Beijing using the developed LP-LIF instrument on several occasions. The results confirmed the instrument's stable operation and demonstrated its excellent adaptability. The developed LP-LIF system provides a highly sensitive and precise tool for quantifying atmospheric oxidizing capacity and further exploring the causes of secondary pollution.
Read full abstract