An isocratic reversed-phase high-performance liquid chromatographic method with ultraviolet detection at 205 nm has been validated for the determination of indinavir, ritonavir and lopinavir (ABT 378) in human plasma. The ritonavir analogue A-86093.0 was used as internal standard. Good chromatographic separation was achieved using a stainless steel column packed with 5 μm Phenomenex phenyl hexyl material operated at 40 °C, and a mobile phase consisting of acetonitrile–10 m M potassium phosphate buffer (50:50, v/v). The calibration curve for indinavir was linear over the range of 50 to1000 μg/l while the ritonavir and lopinavir calibration curves were linear over the range of 100 to 15000 μg/l. The lower limit of quantitations for indinavir, ritonavir and lopinavir were 50, 100 and 100 μg/l, respectively, using 500 μl of human plasma. The validation data showed that the assay is sensitive, specific and reproducible for determination of indinavir, ritonavir and lopinavir. This method is being used in a therapeutic drug monitoring service to quantitate these therapeutic agents in patients infected with human immunodeficiency virus.