A racetrack field asymmetric waveform ion mobility spectrometry (r-FAIMS) device, which consists of both cylindrical FAIMS (c-FAIMS) and planar FAIMS (p-FAIMS) sections with a 1 mm gap width, was developed and applied for high-resolution and high-sensitivity exploration of conformational diversity for peptides. The optimal operating conditions of r-FAIMS were systemically studied, and the performance of the fully optimized r-FAIMS was compared to a previously developed p-FAIMS in detail by using pure nitrogen as the FAIMS carrier gas. Relying on the ion focusing effect in the c-FAIMS section, the intensity of the FAIMS spectrum for doubly charged bradykinin ions acquired by using r-FAIMS is ∼8.5-fold higher than that acquired by using p-FAIMS under the same resolving power/resolution condition, implying about an order of magnitude better sensitivity of r-FAIMS. In addition, the peak separation resolution of r-FAIMS was ∼1.70-fold higher than p-FAIMS under a similar sensitivity condition for doubly charged bradykinin ions. Due to a reduced gap width of the newly designed r-FAIMS (1 mm) as compared to the previously developed p-FAIMS (1.88 mm), r-FAIMS can operate at a much higher separation field with a similar FAIMS dispersion voltage (DV) to gain significantly higher resolving power. For triply charged syntide 2 ions, the resolving power of r-FAIMS can easily exceed 120 at -3.5 kV DV by using pure nitrogen as the FAIMS carrier gas as compared to 44.2 resolving power obtained by using p-FAIMS at -4.0 kV DV. All of the experimental results have confirmed that r-FAIMS can perform structural characterization of biomolecules with both high resolution and high sensitivity.