In this work, the procedure for direct atomic emission spectrometry (AES) determination of fluorine impurities in carbon materials was proposed. Two types of simple radiation sources with direct solid (powdered) sample introduction were studied: direct current arc (25A) and high-power low voltage spark (300 V, of 2348 μF). For each source, standard graphite extra high purity electrodes and argon discharge atmosphere were used. Red spectral region approach to non-metal AES determination was employed with analytical line F I 685.603 nm. The spectra were registered by compact tunable fiber-optic spectrometer with linear complementary metal oxide semiconductor array (CMOS) sensor and short-focus lens. Special measures were taken to counter red spectral region interferences, such as CN molecular bands and thermal radiation. The excitation of F in both types of radiation source was studied. A set of artificial graphite and NaF based standards was prepared for calibration and evaluated by two independent approaches: indirect dc arc AES and water extraction photometry. Limit of detection was determined for dc arc source was 5 μg/g, and for spark 20 μg/g of F. The whole cycle of measurement with each type of radiation source takes 10 min. As the power of dc arc is limited by generator capabilities, it is possible that even lower limit of detection is achievable with higher discharge current.