With the introduction of the MR-Linac, online adaptation becomes reality in the clinic. Real-time adaptation is introduced to account for the anatomic changes during the process of online adaptation and RT delivery. In the abdomen, these changes include motions due to respiration, peristalsis and unpredictable gas filling of the gastrointestinal (GI) organs. While real-time motion tracking, a form of real-time adaptation, is being developed to address the respiratory and/or peristaltic motions, the intrafractional change of the gas cavities and its impact on dose delivery has not been investigated. In this work, we study the change of gas volumes in a time frame similar to the duration of online adaptation and fraction dose delivery and its dosimetric impacts during MRI-guided adaptive RT (MRgART). Time-sequenced abdominal MRI sets (3D T2 and T1, or motion average MRI from 4DMRI) separated by 15-60 minutes (mimicking the duration of MRgART) from 9 abdominal cancer patients were analyzed. For each set, the pancreatic head, duodenum, bowels, and stomach were delineated. Gas volumes were contoured separately and justified with corresponding CT. An IMRT plan with a prescription of 50.4 Gy for 28 fractions to the target (the pancreatic head) was created and calculated under magnetic field on the first MR. The electron density was forced to 0.25 for gas and 1 otherwise. The plan was recalculated on the second MR. The effects on dose volume histogram (DVH) parameters due to gas volume changes for a single fraction are reported. The GI gas cavity volumes can change substantially in the duration of 15-60 min. Gas cavities of the stomach and duodenum were found to vary, on average, by 38 and 3.5 cc, respectively, for the cases studied. Changes in selected DVH parameters resulting from these (intrafractional) gas cavity changes are tabulated for a single fraction. The IMRT plans generated on the pancreatic head showed a medium correlation between the stomach gas volume and the maximum dose to the target (|r| = 0.3) and a strong correlation to the target V(97%) (|r| = 0.7) and to the duodenum (|r| = 0.7) is reported. There was a low to medium correlation (|r| = 0.2 – 0.6) between the duodenum gas volume change and the observed DVH parameters despite the comparably smaller volume change. Intrafractional gas volumes vary rapidly within GI structures causing anatomical changes near the target. Such changes can result in a negative impact on the dose delivery, even with online adaptation. These results demonstrate the need to consider real-time adaptation as a means to address the effects of bowel gas variation during MRgART.Abstract 3754; Table 1Max Target Dose (cGy)Min Target Dose (cGy)Target V(97%) (%)Max Duodenum Dose (cGy)Average Difference1.8324.85.4Max Difference5.2931414Min Difference0.00.20.00.1 Open table in a new tab