PurposeA 2-m axial field-of-view, total-body PET/CT scanner (uEXPLORER) has been recently developed to provide total-body coverage and ultra-high sensitivity, which together, enables opportunities for in vivo time-activity curve (TAC) measurement of all investigated organs simultaneously with high temporal resolution. This study aims at quantifying the cumulated activity and patient dose of 2-[F-18]fluoro-2-deoxy-D-glucose (F-18 FDG ) imaging by using delayed time-activity curves (TACs), measured out to 8-h post-injection, for different organs so that the comparison between quantifying approaches using short-time method (up to 75 min post-injection) or long-time method (up to 8 h post-injection) could be performed.MethodsOrgan TACs of 10 healthy volunteers were collected using total-body PET/CT in 4 periods after the intravenous injection of F-18 FDG. The 8-h post-injection TACs of 6 source organs were fitted using a spline method (based on Origin (version 8.1)). To compare with cumulated activity estimated from spline-fitted curves, the cumulated activity estimated from multi-exponential curve was also calculated. Exponential curve was fitted with shorter series of data consistent with clinical procedure and previous dosimetry works. An 8-h dynamic bladder wall dose model considering 2 voiding were employed to illustrate the differences in bladder wall dose caused by the different measurement durations. Organ absorbed doses were further estimated using Medical Internal Radiation Dose (MIRD) method and voxel phantoms.ResultsA short-time measurement could lead to significant bias in estimated cumulated activity for liver compared with long-time-measured spline fitted method, and the differences of cumulated activity were 18.38% on average. For the myocardium, the estimated cumulated activity difference was not statistically significant due to large variation in metabolism among individuals. The average residence time differences of brain, heart, kidney, liver, and lungs were 8.38%, 15.13%, 25.02%, 23.94%, and 16.50% between short-time and long-time methods. Regarding effective dose, the maximum differences of residence time between long-time-measured spline fitted curve and short-time-measured multi-exponential fitted curve was 9.93%. When using spline method, the bladder revealed the most difference in the effective dose among all the investigated organs with a bias up to 21.18%. The bladder wall dose calculated using a long-time dynamic model was 13.79% larger than the two-voiding dynamic model, and at least 50.17% lower than previous studies based on fixed bladder content volume.ConclusionsLong-time measurement of multi-organ TACs with high temporal resolution enabled by a total-body PET/CT demonstrated that the clinical procedure with 20 min PET scan at 1 h after injection could be used for retrospective dosimetry analysis in most organs. As the bladder content contributed the most to the effective dose, a long-time dynamic model was recommended for the bladder wall dose estimation.