Simultaneous elemental detection of F and Cl offers quantitation of fluorinated and chlorinated compounds and their transformation products without compound-specific standards. Despite wide-ranging applications, this capability has been hindered by fundamental and technical shortcomings of current inductively coupled plasma (ICP)-MS methods in ion formation and isobaric interference elimination. These hurdles are alleviated here via a chemical ionization method. Fluorine and chlorine in analytes are first converted to HF and HCl by an ICP with post-plasma recombination and subsequently react with barium-containing ions supplied by a nanospray, yielding BaF+ and BaCl+ as elemental reporter ions. Notably, the method is readily interfaced to an Orbitrap MS which eliminates isobaric interferences at resolving powers as low as 35,000, far greater than that of current ICP-MS instruments. Moreover, the instrument is easily reverted to the ESI-MS mode for complementary molecular characterization. To demonstrate analytical capabilities, a workflow for rapid quantitation of compounds separated by reversed-phase liquid chromatography is developed using a species-independent calibration. The independent F and Cl measurements agree with each other, providing recoveries of >90% and LODs of 8-12 pmol Cl and 5-12 pmol F on the column. The workflow along with LC-ESI-MS on the same instrument is then applied to identify and quantify in-vitro drug metabolites, yielding total drug-related material recoveries of >80% and quantitation of minor metabolites summing to 8% of the total drug-related compounds. These results highlight the strengths of simultaneous F and Cl speciation for rapid quantitation with applications in early drug development.