To report on the failure rates of an expandable interbody spacer, used for spinal instrumented transforaminal lumbar interbody fusion (TLIF) or posterior thoracolumbar interbody fusion (PLIF). In this retrospective, single-center observational cohort study we reviewed consecutive PLIF/TLIF patients using the Catalyft™ PL and PL40 expandable titanium interbody implant (Medtronic, Minneapolis (USA)) between 07/2022 and 11/2023. We recorded patient demographics, surgical parameters, adverse events, radiological parameters and clinical outcomes according to the MacNab criteria. We identified 53 patients (mean age 68.6 ± 11.5 years; 49.1% female; mean BMI 25.7 ± 4.4kg/m2), in which 92 Catalyft™ spacers were implanted, mostly at L4/5 (n = 35; 38.0%) and L5/S1 (n = 31; 33.7%) for degenerative (n = 44; 83.0%), deformity (n = 7; 13.2%) or other indications (n = 2; 3.8%). We noticed secondary collapse of 20 (21.7%) spacers in 11 patients (20.7%) occurring after a mean time of 4.0 ± 4.1 months postoperatively. Age < 70 years was identified as risk factor for secondary collapse (OR 3.71, 95% CI 1.31-10.5, p = 0.014), but no association was found regarding other patient-specific or surgery-related variables. Pseudarthrosis was evident in seven patients (13.2%), of which four (7.6%) required revision surgery. Clinical outcome at last follow-up (mean 6.4 ± 4.2 months) was excellent/good in 29 (54.7%), fair in 14 (26.4%) and poor in six patients (11.3%). We provide a critical analysis of our series of patients with use of a specific type of expandable interbody spacer. We noticed failure and secondary collapse in an unacceptably high number of implants, some of which required revision surgery.