This article presents a three-axis inductive tactile force sensor utilizing a magnetorheological elastomer (MRE) marker for inducing inductance change in square-shaped coils, in response to an input force. The sensor consists of copper coils on a printed circuit board (PCB) and a patterned elastomer layer between the marker and coils. The proposed configuration of the coils reduces the mechanical crosstalk between the applied three-axis forces, thus decoupling the force axes. Finite-element method (FEM)-based analyses are carried out to optimize the sensing gap between the coils and the marker. To observe the effect of the elastomer material on the sensor performance, two types of silicone rubbers, Ecoflex-30 and RTV-528, are used to make the patterned elastomer and the marker. The sensor with Ecoflex-30 elastomer can measure forces up to 25 N in normal and 2.5 N in shear directions, whereas the sensor with RTV-528 elastomer has a higher force range of up to 30 N for normal and 6 N for shear directions but has a lower sensitivity. A resolution of 12.71, 7.96, and 6.61 mN is achieved with Ecoflex-30, whereas for RTV-528, the force resolution is 76.52, 11.7, and 23.98 mN for normal, shear, and angular shear directions, respectively. The hysteresis error for the proposed sensors is 7.2% and 5.3% for Ecoflex-30 and RTV-528 elastomers, respectively. The proposed inductive tactile force sensor can be used in a wide range of robotic surgical applications for sensing multidirectional forces with better resolution and low hysteresis error.