The purpose of this work was to develop and validate a rapid, non-destructive, real-time measurement system for nanoscale zero-valent iron (nZVI) particles determination in sand-packed columns. This aims to provide an approach to analyse for nZVI particle transport and nZVI induced contaminant degradation reactions in 1-dimensional aquifer experiments. The inductive measurement system (IMS) was developed and tested by measuring columns filled with varying portions of nZVI and sandy aquifer material. Method validation was conducted for this innovative approach in accordance with several guidelines in disciplinary areas. The linearity of ZVI concentration and measurement signal was confirmed. Calibration curves were determined to be linear over the range of 0.11–53.50 g/L for nZVI NANOFER STAR, 0.10–81.67 g/L for nZVI BASF 1, 0.12–58.26 g/L for nZVI Carbo-Iron, 0.12–69.70 g/L for nZVI NANOFER 25DS ORM, 0.12–50.69 g/L for nZVI NANOFER 25DS, and 0.21–1613.39 g/L for nZVI Höganäs 1* (dry). This method was determined to be sensitive in nZVI, accurate (101.63 % within-run recovery, 101.85 % between-run recovery covering 76 days), and precise (0.46 % repeatability RSD, 0.85 % intermediate precision RSD covering 76 days). LOQ was determined ranging from 0.10 to 0.21 g/L for nZVI used in experiments and regression through origins was selected as the calibration model. Quality control processes were performed to assure the reliability of the results and the long-term stability. Additional data processing steps allowed to compensate for measurement deviations resulting from different liquid phase compositions, e.g. dispersants used for nZVI suspensions. This method is now successfully validated in single-laboratory conditions.