Transcatheter aortic valve implantation (TAVI) has become a therapeutic treatment for severe symptomatic patients with aortic stenosis. This study aimed to test a novel transcatheter aortic self-expandable bioprosthesis-the ScienCrown system (Lepu Medtech Inc., Beijing, China)-and evaluate the safety of the new device during TAVI. ScienCrown aortic valve implantation was performed on 10 patients. Clinical assessment was performed at baseline, post procedure, and after 1year. Clinical outcomes and adverse events were assessed according to Valvular Academic Research Consortium-3 criteria. The mean age was 75.30 ± 4.78years with a mean Society of Thoracic Surgeons score of 4.64 ± 3.23%. Device success was achieved in all patients (80% transfemoral, 20% transapical). After 1year, there were no deaths, disabling strokes, myocardial infarctions, conversions to surgery, or major procedure-related complications. New pacemaker implantation was required in one patient (10%). ScienCrown implantation resulted in a reduction in mean valve gradient (63.00 ± 18.84 to 9.67 ± 4.97mm Hg, p <0.001) and an increase in effective orifice area (0.57 ± 0.20 to 2.57 ± 0.59cm2, p <0.001) at 1year. Paravalvular leak was absent in 9 patients (90%), and there was a trace in one patient (10%). All patients were in New York Heart Association class I to II at a mean follow-up of 1year. The experience showed that ScienCrown transcatheter aortic valve system was safely and successfully implanted for treatment of severe symptomatic aortic stenosis. The newer-generation device affords a stable implantation while providing optimal hemodynamic performance.