Intracardiac echocardiography (ICE) is an alternative to transesophageal echocardiography to guide left atrial appendage occlusion (LAAO). However, 2-dimensional ICE has important limitations that hinder its adoption in routine practice. This prospective multicenter study (NCT04569734) investigated the feasibility and safety of ICE-guided LAAO with Watchman FLX using a novel 3-dimensional ICE catheter (VeriSight Pro, Philips). A total of 100 patients undergoing LAAO with the Watchman FLX device were recruited. All cases were performed under moderate sedation. A simplified 2 orthogonal view ICE imaging protocol was adopted to assess PASS (position, anchoring, sizing, and sealing) criteria. The feasibility endpoint was successful implantation of the FLX device and adequate seal (peri-device leak<5mm) at 45days. The safety endpoint was the incidence of major complications at 7 and 45days. Physician experience and imaging quality scores were collected (1=lowest, 5=highest). The mean age was 77.0±8.1years, and 39 (39%) were women. CHA2DS2-VASc score was 4.8±1.5 and HAS-BLED score was 3.2±1.1. Total procedural time was 54±25minutes. Mean contrast volume was 41±46ml, with no contrast used in 16 patients. The feasibility endpoint was achieved in 95 patients (95%); 4 patients were not implanted due to unsuitable anatomy, and 1 patient had a leak >5mm at 45days. There was 1 periprocedural stroke and 2 gastrointestinal bleeds requiring transfusion at 7days with no other major complications. Physicians' rating of overall ICE imaging quality was 4.6±0.6. The novel 3-dimensional ICE VeriSight Pro Probe can safely and effectively guide LAAO with Watchman FLX using a simplified imaging protocol.