Subcutaneous implantable cardioverter-defibrillators (S-ICDs) have been shown to be non-inferior to transvenous ICDs in the prevention of sudden cardiac death (SCD), but there is still alack of evidence from clinical trials in China. We investigated whether S‑ICD implantation in the Chinese population is safe and feasible and should be promoted in the future. Consecutive patients undergoing S‑ICD implantation at our center were enrolled in this retrospective study. Data were collected within the median follow-up period of 554days. Data concerning patient selection, implantation procedures, complications, and episodes of shock were analyzed. In total, 70.2% of all 47patients (median age = 39years) were included for secondary prevention of SCD with different etiologies. Vector screening showed that 98% of patients were with > 1 appropriate vector in all postures. An intraoperative defibrillation test was not performed on six patients because of the high risk of disease deterioration, while all episodes of ventricular fibrillation induced post implantation were terminated by one shock. As expected, no severe complications (e.g., infection and device-related complications) were observed, except for one case of delayed healing of the incision. Overall, 15patients (31.9%) experienced appropriate shocks (AS) with all episodes terminated by one shock. Two patients (4.3%) experienced inappropriate shocks (IAS) due to noise oversensing, resulting in ahigh Kaplan-Meier IAS-free rate of 95.7%. Based on appropriate patient selection and standardized implantation procedures, this real-world study confirmed the safety and efficacy of S‑ICD in Chinese patients, indicating that it may help to promote the prevention of SCD in China.