Introduction To perform end-to-end testing of conformal radiotherapy (3D-CRT), and various Intensity modulated radiotherapy (IMRT) techniques for Addington, Greys and Inkosi Albert Luthuli central hospital (IALCH) using a Varian Eclipse treatment planning system (TPS) and Elekta Mosaic TPS with Varian and Elekta linear accelerator delivery systems. Materials and methods A MatrixX with MULTICube was CT scanned, transferred to TPS, and treatment plans were created for each treatment technique. A virtual water phantom was also created in the TPS and the same plans were generated. The plans were transferred, verified and delivered on the Linacs. The planned dose was then compared with the measured dose in both phantoms. Results The maximum dose (D2) difference was found to be −2.27 ± 0.60%, −1.35 ± 0.51%, −3.68 ± 0.31% and −4.67 ± 0.56% for 3D-CRT, step-and-shoot IMRT (ssIMRT), dynamic IMRT (dIMRT) and intensity modulated arc therapy (IMAT) respectively with the MatrixX. The dose difference in water measured at an isocenter with a Farmer ionization chamber (FC65-G, IBA Dosimetry Group, Germany) was found to be −0.19 ± 0.41%, 0.02 ± 0.47%, 0.08 ± 0.34% and −0.27 ± 0.44% for 3D-CRT, ssIMRT, dIMRT and IMAT respectively. The MatrixX detectors sampling time and 7.6 mm resolution contributed to high dose differences in dIMRT and IMAT techniques. Conclusion End-to-end testing at our institutions was carried out successfully for a range of techniques. The measured and planned doses for both MatrixX and water measurements were comparable and acceptable.