It is difficult to predict the gas transfer performance in oxygenators theoretically. Because there have not been a method which could calculate nonlinear blood-gas reactions and dealt blood and gas layer simultaneously. We have been trying to calculate blood-gas transfer with a CFD method involving nonlinear hemoglobin-gas reactions. In this study, we analyzed distribution of gaseous concentrations and gas transfer performance of a partial oxygenator model. Referenced whole oxy-genator has a rectangular formed hollow fiber bundle. Partial oxygenator model had a part of blood flow section and the same length of primary blood flow direction (30 mm) of whole oxygenator's fiber bundle, and was made of a blood layer, gas layers and staggered cylindrical hollow fibers (225 μm outer diameter). Our method adopted O2 and CO2 reaction models derived on nonlinear relationship equations for source terms in mass transfer equations between each concentration and reaction source. In a result, a state of nonlinear distributions could be observed under Reynolds number=3.9. If a blood flow section of this model converted into that of whole oxygenator in condition of uniform blood flow, 0, and CO, transfer rates were 71.1 mL/min and 80.0 mL/min, respectively at 1 UL/min gas flow, and 71.1 mL/min and 83.7 mL/min, respectively at 3 L/min gas flow in case of 1 L/min blood flow. This calculated results showed an enough performance with reality.