Urine albumin (UA) is an important biomarker of chronic kidney disease. Current in vitro diagnostic medical devices (IVD-MDs) for measuring UA are not standardized, and median results among IVD-MDs differ by approximately 45%. Since fixed decision values are used to interpret UA, higher-order reference measurement procedures (RMPs) are needed for metrological traceability. Three candidate liquid chromatography-tandem mass spectrometry RMPs have been developed for UA. Eight single-donation human urine samples were measured by 3 candidate RMPs. Results were compared using t-test and variance component analysis. The mean results for each urine sample from each RMP laboratory were not statistically different from the overall mean value by t-test. The median total CV including contributions from bias and imprecision among the 3 RMP laboratories was 6.23% using variance component analysis for each sample. The allowable bias to the RMP for an end-user IVD-MD was ≦9.0% or ≦3.0% based on the desirable or optimal total allowable error of 30% or 24%, respectively. A maximum allowable standard uncertainty for an RMP result was determined to be 4.3% or 3.3% for desirable or optimal performance, respectively. The standard uncertainties for all of the RMP laboratories meet the desirable and optimal standard uncertainty specifications. The candidate RMPs for UA in these 3 laboratories have suitable agreement of results and uncertainties for use as higher-order RMPs in the metrological traceability of end-user IVD-MDs for measuring UA.