BackgroundWe compared two Hb A2 methods, high pressure liquid chromatography and capillary zone electrophoresis, at two occasions. In 2014 the methods showed good agreement, while in 2020 HPLC results were clearly higher than CE results. This finding prompted us to investigate the external quality assessment (EQA) outcome and our total patient results obtained by high pressure liquid chromatography over several years. MethodsThe methods compared were Bio-Rad Variant II Beta-Thal Short Program (HPLC) and Sebia Capilllarys Flex Piercing (CE). ResultsOur annual patient results obtained by HPLC increased significantly from 2014 to 2020. A similar trend was also seen in our EQA results. When patient results were grouped according to different reagent lots it became evident that method comparisons might be severely affected by lot-to-lot variation. We also noticed that samples analyzed with the HPLC method following a sample containing a high proportion of Hb E where prone to give falsely increased Hb A2 results. ConclusionStandardization of the measurement of Hb A2 is urgently needed. Furthermore, the lot-to-lot variation must be minimized. While waiting for these improvements each laboratory ought to repeatedly evaluate their distribution of patient Hb A2 results.