Large variability exists among different growth hormone (GH) assays owing to differences in calibration, antibody specificity, isoform recognition, and interference from GH binding protein (GHBP). The GH receptor antagonist Pegvisomant presents a new challenge because Pegvisomant interferes with many GH assays. A recent consensus conference established criteria for standardization and evaluation of GH assays. Following consensus recommendations, we developed a new GH assay on an automated analyzer (IDS-iSYS, Immunodiagnostic Systems). A monoclonal antibody not cross-reacting with Pegvisomant was combined with a monoclonal antibody specific for 22-kD GH. Isoform specificity and interference from GHBP was tested and compared to that seen in 2 existing automated GH assays (Siemens Immulite, Diasorin Liaison). We also compared GH concentrations measured by the 3 assays for healthy volunteers and patients with acromegaly receiving different treatments. Using the iSYS assay, we also established nadir GH values during oral glucose load and analyzed changes in endogenous GH during Pegvisomant treatment. Analytical and functional sensitivities were 0.01 μg/L and 0.04 μg/L, with a dynamic range from 0.04 to 100 μg/L. Intraassay CVs were 2%-4%, whereas interassay CVs were 5%-7% at GH concentrations between 1.7 and 27.5 μg/L. The assay was specific for 22-kD GH and not affected by GHBP. The presence of Pegvisomant, which leads to a negative bias on the Immulite and dramatic overestimation of GH on the Liaison, had no impact on the iSYS GH assay. The new assay fulfils recent consensus recommendations and presents a useful new tool for reliable measurement of GH.