The most severe interferences in atomic absorption spectrometry are caused by the presence of anions when they are in different concentrations in the samples and in the calibration solutions. The analyte addition technique or matrix matching calibration can be employed to minimize or compensate the non-spectral interferences, but they are time consuming or difficult to be carried out. The use of chemical modifiers usually allows higher pyrolysis temperatures and consequently the removal of components of the sample matrix, equalizing the analyte signal in the sample and in the calibration solution. In this work, a mixture of Ir and Rh is proposed as permanent modifier to determine As, Cd and Pb in diluted hydrochloric, sulfuric and phosphoric acids and in ethanol and methanol by electrothermal atomic absorption spectrometry (ET AAS) with calibration against 1% v/v nitric acid aqueous solutions. The performance of the proposed permanent modifier was compared to that of Pd plus Mg nitrates in solution. Better recoveries, low background levels and faster analysis were obtained with the permanent modifier. The permanent modifier was also successfully employed for the determination of As, Cd and Pb in different concentrations of sulfuric and hydrochloric acids. For the phosphoric acid, the proposed modifier was only efficient for acid concentrations up to 2% v/v for As and up to 5% v/v for Cd and Pb. The precision, expressed as the relative standard deviation ( n=3), was lower than 10%, for all samples, including ethanol and methanol.