Projectile X-ray detection (PXD) has been studied to estimate the potential of this ion identification technique in accelerator mass spectrometry (AMS). Limits and possible applications of PXD are discussed and compared to standard isobar identification techniques with gas ionisation chambers. The isobar discrimination capability of a gas ionisation chamber depends on mass, nuclear charge and energy of the ions to be identified. At about 1.5 MeV/amu, suppression factors higher than 10 4 and efficiencies close to 1 have been achieved for ions of masses up to 40 amu. The isobar discrimination capability of PXD is nearly independent of the ion energy, and suppression factors of 10 to 100 are possible. The detection efficiency depends on the ion-target combination, the detector arrangement and the energy of the projectiles. Our results show, that PXD can compete only at lower energies or higher masses, for which gas counters give insufficient isobar discrimination. For the PXD method detection limits on the order of 10 −12 for 36 Cl Cl , 10 −8 for 59Ni Ni and 10 −7 for 126Sn Sn were reached.