Endoscopic ultrasound (EUS) aspiration needles are single-use devices. However, in many centers, because of cost-constraints, these devices are reused multiple times. We studied microbiological contamination and bioburden on reprocessed needles to evaluate whether these devices can be successfully sterilized. We studied 10 EUS needles each of 19G, 22G, and 25G in size, and five 22-G ProCore needles. After initial use, each needle was reprocessed by a standardized protocol. We used standard microbiological cultures, as well as ATP bioluminescence technique to quantify bioburden as relative light units (RLU). We defined significant soil contamination by RLU values >200. We also used extractant fluid to disrupt cell membranes in an attempt to enhance ATP detection. We found culture positivity in 3/34 (8.8%), and detectable bioburden on the exposed surface of 33/35 (94.3%), and inside lumen of 29 (82.9%) reprocessed FNA needles. Significant bioburden was found in three (8.6%) and two (5.7%) needles on the surface and lumen, respectively. We found that use of extractant fluid enhanced detection of bioburden. Larger (19G) needles had higher surface contamination (P=0.016), but there was no relation of luminal contamination with needle diameter (P=0.138). Sheath design and presence of side bevel did not influence extent of contamination. There was significant correlation between the surface and intraluminal bioburden (P<0.001). There is significant bioburden in reprocessed EUS needles; standard microbiological cultures have low sensitivity for detection of needle contamination. We have provided objective evidence for the futility of reprocessing attempts, and practice of EUS needle reuse should be discontinued.