BackgroundThe trans-radial approach for cardiac catheterization led to an increasing adoption of 5 French (F) catheters. We aim to evaluate reliability and reproducibility of coronary physiology assessment performed with 5F guiding catheter (GC). MethodsPhysiological measurements were performed in a coronary flow simulator, which provides two pulsatile flows, the baseline and hyperaemic flows. Two screws, positioned proximally and distally to the distal sensor of a pressure-temperature guidewire, were used to determine various combinations of stenoses and distal obstructions, simulating different pathophysiological conditions. For each setting, 5 measurements of fractional flow reserve (FFR), coronary flow reserve (CFR) and index of microvascular resistance (IMR) were performed with 6F and 5F GCs. ResultsA total amount of 190 measurements were performed, 95 with 6F GC and 95 with 5F GC. Minimal differences between 6F and 5F GCs were detected for FFR [0.91 (IQR: 0.87–0.94) and 0.87 (IQR: 0.82–0.92) respectively, p < 0.001] and IMR (16.5 ± 8.8 and 15.4 ± 8.3 respectively, p = 0.001). Mean CFR was comparable between 6F and 5F GCs (3.6 ± 1.1 and 3.5 ± 0.7 respectively, p = 0.38). Misclassification rates were 1.0 %, 1.0 % and 0 % for FFR, CFR and IMR, respectively. According to Passing-Bablok analysis, an excellent agreement between 6F and 5F GCs was demonstrated for FFR and IMR, and a modest agreement for CFR. All measurements with 5F GC showed high reproducibility. ConclusionsIn our in-vitro model, a complete physiological assessment including FFR, CFR and IMR resulted substantially comparable between 6F and 5F GCs. Further in-vivo analysis is required to support these findings.
Read full abstract