Abstract Flow injection on-line sorption preconcentration and separation in a knotted reactor (KR) was coupled to cold vapor atomic fluorescence spectrometry for the determination of trace mercury in mineral water. Mercury was preconcentrated by on-line formation of mercury diethyldithiocarbamate complex (Hg-DDTC) and absorption of the resulting neutral complex on the inner walls of a knotted reactor. A 20% (v/v) HNO 3 solution heated by Electromagnetic induction heating technique was used as eluent to remove the absorbed Hg-DDTC from the KR, and then the vapor mercury generated by mixing the resulting solution and KBH 4 was determined on-line by cold vapor atomic fluorescence spectrometry. The 20% HNO 3 was used as both the efficient eluent and the required acidic medium for subsequent mercury vapor generation. Using 20% HNO 3 instead of conventional organic solvent as eluent, the proposed method is simple, easy operational and environmental friendly. Under the optimal experimental conditions, the sample throughput was approximately 30 per hour with the enhancement factor of 35. The detection limit of mercury was 2.0 ng l −1 . The precision (RSD, n = 11) was 2.2% at the 0.1 μg l −1 Hg 2+ level.