Volar locking plates designed for far distal radius fracture fixation can have a significant hardware removal rate and risk of tendon rupture.Plate design has a role in the rate of complications. This study assessed the hardware removal and tendon rupture rate of the Acu-Loc 2 volar distal radius (VDR) plate often used in the treatment of far distal radial fractures. We searched our electronic healthcare records system for all patients who had undergone fixation with an Acu-Loc 2 VDR plate (Acumed, Hillsboro, OR, USA) at a tertiary center between January 2017 and December 2021. Patients were excluded if their follow-up time was less than one year or if they could not be contacted by telephone follow-up. Pre-operative radiographs were examined for fracture classification. Follow-up time was defined as the last contact in the clinic or by telephone. A total of 92 patients underwent an open reduction and internal fixation (ORIF) with an Acu-Loc 2 VDR plate. A total of 85 patients met the inclusion criteria for this study. Our cohort included 33 males (38.8%) and 52 females (61.2%). The mean age was 50 years. Twenty-seven fractures (31.0%) were extra-articular, and 60 fractures (69.0%) were intra-articular. The mean follow-up time for the patients was 593.3 days (range 369 to 1185 days).Four patients (4.7%) had their hardware removed. Three (3.5%) patients underwent removal due to tendon irritation and one patient (1.2%) due to a peri-prosthetic fracture around the plate. There were no tendon ruptures recorded. The Acu-Loc 2 VDR plate had a low medium-term hardware removal rate and no tendon ruptures. These rates are lower than would be expected when compared with other far distal plate designs.