IntroductionDetermination of folate insufficiency is of considerable interest given its importance in fetal development and red blood cell formation; however, access to blood tests may be limited due to the requirement for phlebotomy as well as controlled temperature shipping of blood specimens to laboratories for testing due to the inherent instability of folate and its vitamers. MethodsAn LC-MS/MS test was developed and validated for the measurement of 5-methyltetrahydrofolate (5MTHF) in dried plasma specimens collected from fingerstick blood using a laminar flow blood separation device, as well as liquid venous plasma for comparison. Two pre-analytical factors investigated influencing the measurement of 5MTHF in dried plasma were hemolysis of the fingerstick blood during collection and storage/shipment of the dried plasma. ResultsAlthough observed infrequently, hemolysis >10 % resulted in elevated 5MTHF measurements, but hemolysis >1 % resulted in elevated chloride measurements, which were necessary to normalize 5MTHF measurements for variation in volume of dried plasma specimens. Stability of 5MTHF was improved in dried plasma relative to liquid plasma at ambient temperatures, but not sufficiently to allow for uncontrolled temperature shipping despite controlling for humidity and light exposure. Shipping studies emulating ISTA procedure 7D were conducted with a reusable cold packaging solution. The packaging failed to stabilize 5MTHF in dried plasma specimens during a 2-day summer shipping evaluation, but did provide sufficient temperature control to stabilize 5MTHF during the overnight shipping evaluation. ConclusionOur studies provide boundary conditions with respect to hemolysis, storage, and shipping for successful analysis of 5MTHF from dried plasma specimens.