A new analytical procedure for the direct determination of metal impurities in aluminium oxide by ultrasonic slurry sampling–electrothermal vaporization–inductively coupled plasma mass spectrometry (ETV–ICP-MS) is reported. The elements studied include Cr, Cu, Ga, Fe, Mg, Mn, Na, V and Zn. Two samples of fine alumina powder (including the certified reference material NIST SRM 699) were used in this study without any additional pre-treatment. Analytical results of the external calibration using aqueous standards, the external calibration using matched matrix standards and the standard additions method are presented and compared. The most appropriate (average accuracy of 97±5.5% for sample NIST SRM 699) was the standard additions procedure of calibration. An acceptable agreement was found between the results of the slurry sampling procedure and the certified or reference values, and the repeatability of measurements was always better than 13.6%. Some other analytical figures-of-merit,e.g., limits of detection, are also presented. The absolute limits of detection (for the analysis of 15 µl samples) ranged from 0.8 to 8.3 pg. The speed and multi-element analytical potential of this technique seem to be very advantageous.